Question: Q. 83: To enhance the sensitivity, an Ammeter is to be designed with two kinds of graduation on its ...

Q. 83: To enhance the sensitivity, an Ammeter is to be designed with two kinds of graduation on its scale - 0 to 10 A and 0 to 1 A. For that a galvanometer of resistance 50 Ω\Omega and full scale deflection current 1 mA was used along with two resistances R1R_1 and R2R_2 as shown. Either of T1T_1 or T2T_2 is to be used as negative terminal of the Ammeter.

(a) When measuring a current of the order of 0.1 A, which shall be used as negative terminal – T1T_1 or T2T_2?

(b) Find the values of R1R_1 and R2R_2.

Answer

(a) T2T_2

(b) R10.00500ΩR_1 \approx 0.00500 \, \Omega, R20.0450ΩR_2 \approx 0.0450 \, \Omega

Explanation

Solution

The problem describes the design of a multi-range ammeter using a galvanometer and two resistances, R1R_1 and R2R_2. The ammeter has two ranges: 0 to 10 A and 0 to 1 A. We are given the galvanometer's resistance (RgR_g) and full-scale deflection current (IgI_g).

Given values:

Galvanometer resistance, Rg=50ΩR_g = 50 \, \Omega

Full-scale deflection current of galvanometer, Ig=1mA=1×103AI_g = 1 \, \text{mA} = 1 \times 10^{-3} \, \text{A}

Circuit Interpretation:

The diagram shows a common positive terminal ('+') and two selectable negative terminals (T1T_1 and T2T_2). This is a common configuration for a multi-range ammeter where the galvanometer is in parallel with a shunt resistance, and the shunt resistance changes depending on the selected range.

Let's analyze the shunt resistance for each range:

  1. When T1T_1 is used as the negative terminal:
    The total current enters at '+' and exits at T1T_1.
    From the diagram, the galvanometer (G) is in parallel with the resistance R1R_1.
    So, the shunt resistance for this range is Rsh1=R1R_{sh1} = R_1.

  2. When T2T_2 is used as the negative terminal:
    The total current enters at '+' and exits at T2T_2.
    From the diagram, the galvanometer (G) is in parallel with the series combination of R1R_1 and R2R_2.
    So, the shunt resistance for this range is Rsh2=R1+R2R_{sh2} = R_1 + R_2.

Relating ranges to terminals:

For an ammeter, a smaller shunt resistance allows a larger current to be measured (higher range), as more current is bypassed from the galvanometer. Conversely, a larger shunt resistance allows a smaller current to be measured (lower range).

Since R1<(R1+R2)R_1 < (R_1 + R_2), the shunt resistance R1R_1 corresponds to the higher current range, and the shunt resistance (R1+R2)(R_1 + R_2) corresponds to the lower current range.

Therefore:

  • Using T1T_1 as the negative terminal corresponds to the 0 to 10 A range.
  • Using T2T_2 as the negative terminal corresponds to the 0 to 1 A range.

(a) When measuring a current of the order of 0.1 A, which shall be used as negative terminal – T1T_1 or T2T_2?

A current of 0.1 A is best measured using the 0 to 1 A range for higher accuracy and sensitivity. As determined above, the 0 to 1 A range uses T2T_2 as the negative terminal.

(b) Find the values of R1R_1 and R2R_2.

The principle for an ammeter is that the voltage across the galvanometer and the shunt resistance is the same at full-scale deflection.

IgRg=(ImaxIg)RshI_g R_g = (I_{max} - I_g) R_{sh}

Where ImaxI_{max} is the full-scale current for the specific range.

For the 10 A range (using T1T_1):

Imax1=10AI_{max1} = 10 \, \text{A}

Rsh1=R1R_{sh1} = R_1

IgRg=(Imax1Ig)R1I_g R_g = (I_{max1} - I_g) R_1

(1×103A)×(50Ω)=(10A1×103A)R1(1 \times 10^{-3} \, \text{A}) \times (50 \, \Omega) = (10 \, \text{A} - 1 \times 10^{-3} \, \text{A}) R_1

0.05=(9.999)R10.05 = (9.999) R_1

R1=0.059.999R_1 = \frac{0.05}{9.999}

R10.0050005ΩR_1 \approx 0.0050005 \, \Omega

For the 1 A range (using T2T_2):

Imax2=1AI_{max2} = 1 \, \text{A}

Rsh2=R1+R2R_{sh2} = R_1 + R_2

IgRg=(Imax2Ig)(R1+R2)I_g R_g = (I_{max2} - I_g) (R_1 + R_2)

(1×103A)×(50Ω)=(1A1×103A)(R1+R2)(1 \times 10^{-3} \, \text{A}) \times (50 \, \Omega) = (1 \, \text{A} - 1 \times 10^{-3} \, \text{A}) (R_1 + R_2)

0.05=(0.999)(R1+R2)0.05 = (0.999) (R_1 + R_2)

R1+R2=0.050.999R_1 + R_2 = \frac{0.05}{0.999}

R1+R20.05005005ΩR_1 + R_2 \approx 0.05005005 \, \Omega

Now, we can find R2R_2:

R2=(R1+R2)R1R_2 = (R_1 + R_2) - R_1

R2=0.050.9990.059.999R_2 = \frac{0.05}{0.999} - \frac{0.05}{9.999}

R2=0.05(10.99919.999)R_2 = 0.05 \left( \frac{1}{0.999} - \frac{1}{9.999} \right)

R2=0.05(9.9990.9990.999×9.999)R_2 = 0.05 \left( \frac{9.999 - 0.999}{0.999 \times 9.999} \right)

R2=0.05(99.989001)R_2 = 0.05 \left( \frac{9}{9.989001} \right)

R2=0.459.989001R_2 = \frac{0.45}{9.989001}

R20.04504955ΩR_2 \approx 0.04504955 \, \Omega

Rounding to a reasonable number of significant figures (e.g., 3-4):

R10.00500ΩR_1 \approx 0.00500 \, \Omega

R20.0450ΩR_2 \approx 0.0450 \, \Omega

(a) Answer:
For measuring a small current like 0.1 A, the 0-1 A range is more sensitive and accurate than the 0-10 A range. In this ammeter design, the 0-1 A range is achieved by using T2T_2 as the negative terminal, which places a larger shunt resistance (R1+R2R_1+R_2) in parallel with the galvanometer, allowing less current to bypass the galvanometer for a given total current.

(b) Answer:
The values of R1R_1 and R2R_2 are calculated using the shunt formula Rsh=IgRgImaxIgR_{sh} = \frac{I_g R_g}{I_{max} - I_g}. For the 10 A range, the shunt is R1R_1. For the 1 A range, the shunt is R1+R2R_1+R_2. Solving these two equations simultaneously yields the values for R1R_1 and R2R_2.